relative quantification of gene

Use this category for questions related to various quantitative PCR methods, including real time PCR, reverse transcription (RT)/PCR, qPCR, etc. For general PCR methods, continue to use the DNA/General PCR forum

Moderators: r.rosati, Suzanne

Re: relative quantification of gene

Postby JMG » May 04 2012 11:34 am

The only way to get your cDNA to behave as if it were 1000X more concentrated
(Cq of 35 to a Cq of ~25) would be if your target gene were more abundant within your
cDNA samples.

If you are indeed just looking for a single copy of something ... yes -- it will appear
anywhere from 37 to 48 Cq (depending on reaction efficiency).

You can't force its Cq value to be any earlier unless there are actually more copies of
the target in the reaction (in the cDNA samples themselves).

Using the cDNA as concentrated as possible won't help much -- and may inhibit the
qPCR. So, a rare target is a rare target -- and a late Cq value is the truth.
JMG
PI of Posters
PI of Posters
 
Posts: 608
Joined: Dec 18 2008 10:13 pm

Re: relative quantification of gene

Postby ashu330327 » May 06 2012 12:15 am

Thanks you soo much Sir...!!! :D
ashu330327
technophile
technophile
 
Posts: 41
Joined: Apr 19 2012 9:22 am

Re: relative quantification of gene

Postby ashu330327 » May 10 2012 11:43 am

Dear Sir
Recently I have done relative quantification of five rpoH genes wrt rpoD as a housekeeping gene by qpcr(Roche) using SYBR GREEN I. I have got cp values of rpoh 1=26, rpoh2,3,4,n 5 above 30. These all are single copy gene. 3 ug RNA was taken for cDNA synthesis and was diluted 3 times. final reaction volume was taken 10 u litre for qpcr.Is cp values are acceptable? i repeated the experiments many times by using lesser amount(1 ug) to higher amount 5 ug RNA and diluted it 3 to 5 times but cp values are above 30. Please help me sir

REACTION EFFICIENCY
rpoh1 1.73
rpoh2 1.82
rpoh3 1.83
rpoh4 1.68
rpoh5 1.85
ashu330327
technophile
technophile
 
Posts: 41
Joined: Apr 19 2012 9:22 am

Re: relative quantification of gene

Postby Selvaraj » Jul 19 2012 8:09 am

Single copy of gene does not mean single copy of transcript.There may be thousands of transcripts for any single copy of gene. Generally, 1ul of 5 ng/ul of cDNA per 10 ul reaction is sufficient.
Selvaraj
technician-in-training
technician-in-training
 
Posts: 5
Joined: Dec 07 2006 2:24 am

Re: relative quantification of gene

Postby memari » Aug 01 2012 6:25 pm

memari
supertech
supertech
 
Posts: 81
Joined: Sep 16 2004 1:35 am

Re: relative quantification of gene

Postby mchlbrmn » Aug 02 2012 12:29 am

Hello. I will first say that I am not a qPCR expert, but I do use it sometimes and these are my impressions.
If you want to know if your Cp's over 30 are valid, you should run a series of 1/10 dilutions (in duplicate or triplicate) with a strong positive sample (relative to your other samples). If the efficiency is a perfect 2fold/cycle, the spacing should be 3.3 cycles. It may be less in the concentrated samples if inhibition is delaying the concentrated sample. It may be more in the dilute sample if the PCR is starting to fail. At very high Ct it may become erratic because initial target DNA molecule numbers are so low that random segregation of the molecules can put more copies in one tube than another. Looking at the dilution series you can judge what range of Ct's is working well (for this template at least, but one can't always do everything) and call the Ct's greater than this 'beyond the range of detetection in this experiment', and those Ct's lower than the good range are also not good. You may not achieve a perfect 3.32 in every dilution, but it should not be too far away.

Looking back, I also noticed that there may have been miscommunication involving the difference between a "single copy gene" (=one copy/genome, but many in the initial tube) and a "single copy of a gene" (=one target molecule of a gene present in the tube).

Oh, I just noticed that you listed PCR efficiency values. So, you already did the dilution curves, and you should be able to look at these, if they extend far enough, to see what Cp value your experiment should be valid.

Another thing to consider is if (when using SYBR PCR) the melt curve gives a single peak. If high Cp samples show an extra peak(s) appearing, and the Cp spacing between 1/10 dilutions becomes lower, this indicates that a nonspecific PCR product is appearing, and affecting the Cp.
If a part of the dilution standard curve is not good, when you remove these samples and recalculate the efficiency, it should become better. (So, instead of having Cp's of 26 and 35 for two samples, the Cp's may now be 26 for one sample and greater than 32, or not detectable within the range of the experiment for the other).

Actually, the efficiency does not need to be 2.0 (and the Cp spacing 3.2) as long as it stays consistent (for example 3.0 Cp between every dilution). Your efficiencies listed look OK to me, but I wonder why there are none that are 1.9 or 2.0. Perhaps if the Cp range of the experiment is reduced, as I described, efficiency would improve (maybe in the last dilution the PCR is starting to fail). However, as long as the Cp's are still increasing with the dilutions, even if the spacing is not perfect, the experiment is still quantitative to some extent (I don't know how much), especially if you are looking at very major differences in expression.