Postby gula » Jun 09 2012 11:59 am

Hi!
I have to introduce a new analysis in my clinical lab. The test demands a DNA-concentration (after extraction out of FFPET) of 2 ng/µl confirmed with Nanodrop.
I don't have this instrument. We use the Qubit.
I know that the Nanodrop shows higher dsDNA-concentration than Qubit, especially when mixed with other nuclein acids.
Our samples are purified with automated Qiacube technique. So I think contamination with RNA isn't high.

Has anyone made a direct comparison of Qubit and Nanodrop figures in pure dsDNA eluates?

I've found only few figures in internet, that show a huge difference under 1 ng/µl but in higher range a factor of about 1,4.
I have eluates with about 1-150 ng/µl to work with.

thank you in advance for your inputs!
Gula
gula
technician-in-training
technician-in-training
 
Posts: 10
Joined: Mar 22 2009 8:06 am

Re: Qubit vs. Nanodrop in clinical lab

Postby mchlbrmn » Jun 09 2012 3:20 pm

I've been using a Nanodrop, and find it very convenient to use. Based on numbers I"ve gotten repeating the same sample, I wouldn't trust it's accuracy below 5 or 10 ng/ul. However, I've not tried to do this so I haven't checked out if there are settings, such as the length of time for the reading, that could be adjusted to improve the accuracy, or if there are other Nanodrops more accurate than my model. I usually take readings of between 50-1000 ng/ul, and it doesn't matter if I'm off by 10%. You might check the manufacturer's specifications to start with.
mchlbrmn
ModSquad
ModSquad
 
Posts: 3328
Joined: Dec 13 2005 1:03 pm
Location: Boston, USA

Re: Qubit vs. Nanodrop in clinical lab

Postby gula » Jun 10 2012 2:02 am

Thanks for your answer.

Do you have directly compared results of nanodrop and qubit of DNA-eluates?

The point is, that the test demands the specified concentration of 2 ng/µl - and that has to be confirmed with nanodrop.
I want to find out a factor for calculation. In other words: what gives a 2 ng-nanodrop-DNA eluate measured in Qubit?

gula
gula
technician-in-training
technician-in-training
 
Posts: 10
Joined: Mar 22 2009 8:06 am

Postby mchlbrmn » Jun 10 2012 8:47 pm

I have never used Qubit.
In my experience, Nanodrop doesn't measure 2ng/ul accurately. However, as I explained, I do not know for sure whether or not Nanodrops are capable of doing this.
mchlbrmn
ModSquad
ModSquad
 
Posts: 3328
Joined: Dec 13 2005 1:03 pm
Location: Boston, USA

Re: Qubit vs. Nanodrop in clinical lab

Postby r.rosati » Jun 11 2012 2:29 pm

I agree, by my experience a Nanodrop isn't reliable at 2 ng/ul.
Nor can similar models - at the moment I use a NanoVue (GE) and it's unreliable below 15-20 ng/ul.
r.rosati
ModSquad
ModSquad
 
Posts: 1799
Joined: Nov 04 2002 10:23 am
Location: Brazil

Re: Qubit vs. Nanodrop in clinical lab

Postby gula » Oct 16 2012 11:57 am

After these few months we got both measurement systems now. I made a dilution row of the DNA-eluat (FFPE samples, QIAamp kit), and compared the results of Qubit and Nanodrop. There was no real linearity, but in mean the results of Nanodrop were 2,6 x higher than those of Qubit.