Non-specific binding from plasma in sandwich ELISA

Use this category for questions regarding various immunological methods

Moderator: mdfenko

Non-specific binding from plasma in sandwich ELISA

Postby leasky » Jun 12 2017 9:25 am

Hello everyone,

I am having some issues reducing non-specific binding in my sandwich ELISA when assaying human plasma samples. For some background, we use our ELISA for detecting a tumour marker in human plasma. As a quality control we spike the marker into human plasma (pooled from normal, healthy population) at various concentrations to ensure consistency in the assay. However, a recently purchased batch of human plasma is producing a high concentration of the tumour marker without any spiked in. We know this is non-specific as this particular marker should be close to zero in the normal population. We have never had an issue with this before, but it is concerning now knowing that our assay isn't 100% efficient in eliminating interference.

I have tried several things, such as: changing brands of plates (switching from Nunc Maxisorp to Nunc Medisorp plates has improved things slightly), trying different blocking buffers with the primary ab (Tween-20 seems to do a better blocking job than BSA) and optimising the wash buffer. Increasing NaCl concentration in the wash to 0.5M reduced non-specific binding and going up to 1M NaCl reduced it even further, and without compromising binding in the standard curve. Implementing these changes has reduced non-specific binding by about 75%, but of course I would like it to be a close to 100% as possible!

I would greatly appreciate any advice from anyone who has had similar problems or who has greater expertise than myself.

Thanks in advance :)
leasky
newcomer
newcomer
 
Posts: 2
Joined: Jun 12 2017 8:47 am

Re: Non-specific binding from plasma in sandwich ELISA

Postby relaxin » Jun 13 2017 10:21 am

Is it possible the new batch of human plasma "contaminated" with tumor marker? Did you try another batch of human plasma?
Retired academic researcher. Mention of a specific product does not imply my endorsement of the product. No conflict of interest or guarantee to work on the advice given. Do as I say, not as I do. Not liable to the loss of your valuable samples.
relaxin
PI of Posters
PI of Posters
 
Posts: 7160
Joined: Jan 11 2006 12:40 pm
Location: Mauna Kea

Re: Non-specific binding from plasma in sandwich ELISA

Postby leasky » Jun 15 2017 8:27 am

Hi Relaxin,

Thanks for your comments, it seems unlikely that it is contaminated with the marker as on one lot of plates we get a high concentration and on another lot of plates we do not. This suggests interference only on certain plates. I understand that it is well known that plate manufacturers can struggle with batch to batch consistency, so there doesn't seem to be a lot I can do about that. My main focus is on getting my assay to the point where it can negate those plate inconsistencies (if that's possible).

To answer your second question, we have multiple batches of plasma that seem to give different baselines of the marker depending on which batch of plates is used (sometimes zero, sometimes in the terminally ill range!). Again, suggesting huge plate inconsistencies. I can't blame everything on the plates though, if my assay was perfect I wouldn't seeing such huge variations.
leasky
newcomer
newcomer
 
Posts: 2
Joined: Jun 12 2017 8:47 am

Re: Non-specific binding from plasma in sandwich ELISA

Postby relaxin » Jun 16 2017 11:29 am

It seems to me that you need to try different blocking buffers rather than optimizing washing solutions. A number of blocking buffers are available from companies such as ThermoFisher, "StartingBlock" is known to be a good blocking buffer. Blocking buffer containing casein may be better than that containing BSA. Washing buffer containing high salt should be avoided.
Retired academic researcher. Mention of a specific product does not imply my endorsement of the product. No conflict of interest or guarantee to work on the advice given. Do as I say, not as I do. Not liable to the loss of your valuable samples.
relaxin
PI of Posters
PI of Posters
 
Posts: 7160
Joined: Jan 11 2006 12:40 pm
Location: Mauna Kea