Postby Tunnelrat » Nov 16 2017 10:16 am

Dear all,

I am having an issue with an ELISA for mouse IgG: samples return lower values on the right side of the ELISA plate when compared to the left side of the ELISA plate. The difference is about 50%.

Although it is not our core business, I have done different ELISA's before with good results.

We have done some trouble shooting:
- We tested an in-house developed ELISA as well as 2 different commercial ELISA kits.
- The plate reader was excluded as the source of error.
- Temperature effects were excluded as the source of error.

Could this somehow be caused by the mouse IgG we want to measure, for example instability/denaturation of the protein in contact with air? Or am I doing something wrong?

Best regards,

Marcel
Tunnelrat
newcomer
newcomer
 
Posts: 1
Joined: Nov 16 2017 9:50 am

Re: ELISA trouble

Postby shakeel » Nov 17 2017 4:56 am

False signals can be generated as a result of non-specific binding of IgG present in the sample. Also, many biological samples are made of a complex mixture which consists of many different proteins and salts, these can also reduce the binding capability of the antigen to the antibody. I would also rule out blocking and washing steps since these are well known causes and also exposing the substrate to light. Some interesting information can be found on the following to webpages.

http://www.elisakits.co.uk/general-elisa-faq/
http://www.elisakits.co.uk/trouble-shooting-faq/

Hope this helps.
shakeel
newcomer
newcomer
 
Posts: 2
Joined: Nov 14 2017 4:32 am

Re: ELISA trouble

Postby r.rosati » Nov 17 2017 12:00 pm

shakeel wrote:False signals can be generated as a result of non-specific binding of IgG present in the sample. Also, many biological samples are made of a complex mixture which consists of many different proteins and salts, these can also reduce the binding capability of the antigen to the antibody. I would also rule out blocking and washing steps since these are well known causes and also exposing the substrate to light. Some interesting information can be found on the following to webpages.

http://www.elisakits.co.uk/general-elisa-faq/
http://www.elisakits.co.uk/trouble-shooting-faq/

Hope this helps.


Hello,
as per Forum policy, if you have commercial interest related to the links, please disclose it, so that we can monitor abuse.
r.rosati
ModSquad
ModSquad
 
Posts: 2136
Joined: Nov 04 2002 10:23 am
Location: Brazil

Re: ELISA trouble

Postby shakeel » Nov 21 2017 6:46 am

No I do not have any commercial interest relating to the links I have posted.
shakeel
newcomer
newcomer
 
Posts: 2
Joined: Nov 14 2017 4:32 am

Re: ELISA trouble

Postby r.rosati » Nov 21 2017 8:50 am

shakeel wrote:No I do not have any commercial interest relating to the links I have posted.


I take it that you are not Shakeel Hussain, from Biosupply UK, the mothership company that created elisa-kits.co.uk.
r.rosati
ModSquad
ModSquad
 
Posts: 2136
Joined: Nov 04 2002 10:23 am
Location: Brazil


Return to Immunology and Immunochemistry Methods

Who is online

Users browsing this forum: No registered users and 1 guest